    We are having the analytical balance with 3 kgs capacity, our operating range is up to 01 Kg only. So we are calibrating the balance daily and uncertainty checking with monthly frequency up to operating range only. Is there any regulation to perform the calibrations up to balance range? If so, please advise…

    Is Eccentricity checking [Corner load test] is mandatory for balance calibration?

    We need experts comments

    Thanks and regards
    V.J.V.Prasad
    Quality Assurance Manager
    jyotiprasad_pharma@yahoo.co.in

  2. #Ads
    AskaboutValidation jyopravel is offline

    V.J.V.Prasad
    Quality Assurance Manager
    jyotiprasad_pharma@yahoo.co.in
     

  3. #2

    Default Balance calibration

    Yes jyo,

    There is regulation like NPL guide and OIML say about this and Eccentricity checking [Corner load test] is mandatory for balance calibration, to be performed with 1/3 of its capacity or 50% of its capacity ( refer guideline), frequency given by Balance Supplier.