Hello Guest! Log In or Register now. Lost Password? Click Here to Reset

Ask About Validation

Results 1 to 7 of 7
  1. #1
    Validation Analyst
    Join Date
    Aug 2012
    Posts
    14

    Post freeze-drier cleaning

    can somebody tell me how to apporach to the cleaning validation of the freeze-drier?should I consider all the products that are lyophilized in order to calculate the limit for swab? in this case, have I to consider the surface of the chamber for the calculation? or this type of equipment can be considered as not at direct contact? is there any references to the guidelines that talk about this aspect?than you for your support!

  2. #2
    Validation Expert
    Join Date
    Mar 2012
    Location
    Hyderabad
    Posts
    116

    Default

    since product cannot touch any part of lyophilizer directly, you can consider it as non-contact equipment and enough to clean upto visually clean state.

    Regards
    Veerraju

  3. #3
    Validation Guru
    Join Date
    Oct 2010
    Location
    Kalundborg,Denmark
    Posts
    1,689

    Default

    Professionals often have a misconception and think that the lyophilizer is a non-contact part.
    In reality and as per regulatory compliance the whole lyophilizer is a contact part based on its operating conditions and movement of various trays, rubber stoppering systems and automation with in Lyophilizer.

    If we have a spillage and lyophilize any other batch with out cleaning, the spillage will eventually dry out and due to vaccum conditions the falkes of these previous spills might enter the product under consideration.

    FDA looks keenly into cleaning of such lyophilizers where end products are rendered stable by application of vaccum, heating and or cooling below subzero temperatures.

    One must thoroughly clean and validate a lyophilizer.This is a critical point where the product quality might be compromized.

    Regards
    Durga Prasad
    Internal Consultant,Aseptic Process-NNE Pharmaplan
    www.nnepharmaplan.com
    Denmark

  4. #4
    Validation Guru
    Join Date
    Oct 2010
    Location
    Kalundborg,Denmark
    Posts
    1,689

    Default Is the freeze dryer a contact part?

    Technically no, but items that can come into contact with the product, e.g.aerating gases, result in the freeze dryer being accepted as a contact part.
    Calculation:
    1/1000 of minimum therapeutic dose in unit dose of next product
    ( swab area contamination * FD surface area / no of vials per batch)
    Durga Prasad
    Internal Consultant,Aseptic Process-NNE Pharmaplan
    www.nnepharmaplan.com
    Denmark

  5. #5
    Validation Analyst
    Join Date
    Aug 2012
    Posts
    14

    Default

    thank you for your point of view. when you talk about regulatory compliance do you make reference to a specific guidelines?
    if yes, could you be so kind to tell me the references?
    I am quite sure that the lyo has to be validated from a cleaning point of view, but the
    problem is the limit calculation. if I consider the surfaces of the lyo shelves that could potentially be contamianted I often have very low limits.
    when you talk about the formula:

    1/1000 of minimum therapeutic dose in unit dose of next product
    ( swab area contamination * FD surface area / no of vials per batch)

    what do you mean with FD surface area?
    thank you for your help.


    Quote Originally Posted by DURGA PRASAD View Post
    Professionals often have a misconception and think that the lyophilizer is a non-contact part.
    In reality and as per regulatory compliance the whole lyophilizer is a contact part based on its operating conditions and movement of various trays, rubber stoppering systems and automation with in Lyophilizer.

    If we have a spillage and lyophilize any other batch with out cleaning, the spillage will eventually dry out and due to vaccum conditions the falkes of these previous spills might enter the product under consideration.

    FDA looks keenly into cleaning of such lyophilizers where end products are rendered stable by application of vaccum, heating and or cooling below subzero temperatures.

    One must thoroughly clean and validate a lyophilizer.This is a critical point where the product quality might be compromized.

    Regards

  6. #6
    Validation Guru
    Join Date
    Oct 2010
    Location
    Kalundborg,Denmark
    Posts
    1,689
    Durga Prasad
    Internal Consultant,Aseptic Process-NNE Pharmaplan
    www.nnepharmaplan.com
    Denmark

  7. #7
    Validation Guru
    Join Date
    Oct 2010
    Location
    Kalundborg,Denmark
    Posts
    1,689

    Default

    There is another way to clean lyophilizers.
    First find out if the product of interest is soluble in 70% or 99% of IPA(iso propyl alcohol).
    If it is soluble, you can take a clean wipe and clean the surface area of the Lyophiliser with that Wipe.
    Incase if it is not soluble the last resort is CIP/SIP. Then it is compulsory to perform a CIP/SIP cycle and validate that.
    Regards
    Durga Prasad
    Internal Consultant,Aseptic Process-NNE Pharmaplan
    www.nnepharmaplan.com
    Denmark

Bookmarks

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •  

Get In Touch with AskaboutValidation

© 2006 - 2012 askaboutvalidation. All Rights Reserved
Single Sign On provided by vBSSO