Smile Validation of SW with Equipment

Hi Everybody,

I have some scenarios in my mind and I need to make sure which deliverables are necessary for the scenarios.

Scenario 1)

I will buy a SW for calibration of pipettes in the laboratory (e.g pipette tracker). I will install the SW on a PC which will connect to a server to store the calculated data. Some equipment is also connected to the SW to deliver some values like temperature, air pressure, humidity...
Also one scale (which is stand-alone qualified) will deliver inputs to the software for calculations.


I have read the whole GAMP 5 but it says nothing about eqipment which will run together with a SW.

Which documents do I need for the whole validation procedure ?

Which regulations apply for USA ?

Which regulations apply for EU ?

Thanks all !
Cihan
Reply With Quote
  #2  
Old 12-17-2009, 12:34 PM
nathanbv nathanbv is offline
Validation Analyst
 
Join Date: Sep 2007
Location: Hyderabad
Posts: 14
Send a message via Yahoo to nathanbv
Default

GAMP5 provides general and overarching guidelines on software and automated systems validation.

If you need specific details on laboratory systems validation you can refer to GAMP guideline on Lab systems or GxP testing.

If you need any further details pease let me know.

Vanmeeganathan B
+91 9704500710
Reply With Quote
  #3  
Old 07-19-2010, 09:43 PM
cihan cihan is offline
Validation Analyst
 
Join Date: Dec 2009
Posts: 2
Default any ideas?

No one has ever validated a pipette calibration software?
Reply With Quote
  #4  
Old 07-26-2010, 11:06 AM
alexkennedy alexkennedy is offline
Validation Expert
 
Join Date: Jan 2008
Location: Argyll
Posts: 95
Default Software and Pipette calibration

Cihan

First you should layout a User requirements requirement stating exactly what you want this system to do.
List everything from how accurate you want it to be, to:what training is required to use it.

Once you have developed this URS and subjected it to peer review and company approval, you will be able to develop IQ/OQ/PQ protocols which will validate that your URS requirements are delivered.

Regards
Alex Kennedy
www.validation-online.net
Reply With Quote
Reply

Thread Tools
Display Modes

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off

Forum Jump

All times are GMT. The time now is 08:53 PM.


Powered by vBulletin® Version 3.8.4
Copyright ©2000 - 2010, Jelsoft Enterprises Ltd.